reading purity threads from 2024 and half of it aged badly
reading purity threads from 2024 and half of it aged badly — a position I have arrived at slowly and would like tested. A shoulder that does not baseline-resolve cannot be quantified honestly. You can report it as an unresolved shoulder, which is useful information, or you can develop the method until it resolves.…
System suitability — repeat injections, tailing factor, plate count, RSD on area — is what tells you the instrument was fit for the measurement that day. Without it, the purity figure is unanchored.
System suitability — repeat injections, tailing factor, plate count, RSD on area — is what tells you the instrument was fit for the measurement that d
This is the whole methodological point of the board in one line.
Small fix — 214nm, not 210. It matters for the comparison you are making with the other run.
That is area percent, not mass percent. The trace cannot give you the second one.
Can you post the trace with the axes labelled?