proposal — a flair for integration posts. Comments open, and the wording is genuinely up for discussion. UV response is not uniform across species. At 214nm you are looking at the peptide bond, which is why it is the working wavelength for this class; at 280nm you only see tryptophan, tyrosine and phenylalanine.…
System suitability — repeat injections, tailing factor, plate count, RSD on area — is what tells you the instrument was fit for the measurement that day. Without it, the purity figure is unanchored.
System suitability — repeat injections, tailing factor, plate count, RSD on area — is what tells you the instrument was fit for the measurement that d
Agreed. And the corollary is that the method line on a certificate is not decoration, it is the number’s provenance.
two labs, two gradients, two honest answers
Same method at both labs, or two different gradients?
Small fix — 214nm, not 210. It matters for the comparison you are making with the other run.
[deleted]
Correction: that is a tailing factor, not a plate count. Different diagnostics for different problems.
axis labels or the trace is decoration
axis labels or the trace is decoration
Disagreeing with this bit: that spread is ordinary inter-lab variance, not a disagreement about the material.