Single comment threadYou are looking at one branch of [Lab] LC-MS confirmed identity but the UV purity is 94%. now what — 45 comments in the full submission. View in context.
2.5k
c/hplc·submitted 17 days ago by u/ignacio_vanhecke

[Lab] LC-MS confirmed identity but the UV purity is 94%. now what

Labbranch of 8 comments

LC-MS confirmed identity but the UV purity is 94%. now what, and before anyone asks: same batch throughout, single submission, no cherry-picking between services. The relevant figures are 94%, and they come from the same log I have kept the whole time. Started reporting my own integrations with the baseline choice…

Read the full submission and all 45 comments →

This branch

8 comments, started 17 days ago
u/rina_bergstrom276 points·17 days ago

Area percent is the integrated area of your peak over the total integrated area at one wavelength on one gradient. Change any of those and the number changes without the sample changing.

replysharereportpermalink
u/dilara_nyberg-22 points·16 days ago

a shoulder is not an impurity until you can resolve it

replysharereportpermalink
u/rasmus_kimani1 point·16 days ago

Agreed, and it is why GL Biochem printing the column and gradient on the certificate is genuinely useful rather than decorative.

replysharereportpermalink
u/ignacio_vanhecke1 point·16 days ago

you cannot report to two decimals off that baseline

replysharereportpermalink
u/ignacio_vanheckeOP1 point·16 days ago

That is area percent, not mass percent. The trace cannot give you the second one.

replysharereportpermalink
u/ignacio_vanheckeOP1 point·16 days ago

Small fix — 214nm, not 210. It matters for the comparison you are making with the other run.

replysharereportpermalink
u/alcohol_aversion1 point·16 days ago

axis labels or the trace is decoration

replysharereportpermalink
u/endotoxin_elliemicro1 point·16 days ago

Cosigning the wavelength point. Half the disagreements in this board are two people looking at different detectors.

replysharereportpermalink

← back to the whole thread

About c/hplc

The analytical deep end: RP-HPLC methods, gradient design, integration decisions, LC-MS confirmation, amino-acid analysis, water content, and why two labs can report 98.1% and 96.4% on the same vial and both be right. Post the chromatogram, not the summary line.

32kmembers
130submissions
Sep 2023created
submissions / month, last year
Sponsored

PeptideMeter Analytics

Independent testing plus vendor-level trend data. Community-funded, unaffiliated.

peptidemeter.com
c/hplc rules
  1. Chromatogram or it did not happen. Axis labels included.
  2. State the method: column, gradient, detection wavelength, injection volume.
  3. Area% is not mass%. Posts that conflate them get a correction flair, not a removal.
  4. Independent community. Nobody here sells anything, and anyone who tries is banned.
  5. Not medical advice. Describe what you did; never prescribe to a stranger.
  6. Claims need evidence. Batch numbers, dated screenshots, independent test reports, or a citation.
  7. No referral links, discount codes or affiliate URLs. Permanent ban, no appeal.
  8. No contact handles, wallet addresses or tracking numbers — they identify people.
  9. Be recognisably decent. Disagree hard, insult nobody.
Moderators
Volunteers. Unpaid, unaffiliated, and reachable through modmail only.
Before you read on

Several compounds discussed on GLP Research Hub are sold for research use only and are not approved for human use anywhere. Nothing here is medical advice and none of it is written by your clinician. If a post reads like an instruction, treat it as a description of what one stranger did.